3 years 8 months ago #5687
by mw0uzo
Replied by mw0uzo on topic Gradual count increase at FSM19
Use one of those filtered mains leads with surge protections etc? Put a few 100nF caps across the supply output?
The following user(s) said Thank You: FSM19
Please Log in or Create an account to join the conversation.
3 years 8 months ago - 3 years 7 months ago #5689
by FSM19
Replied by FSM19 on topic Gradual count increase at FSM19
I'm going offline for a day or so while I design and build a new linear power supply with good in and out filtering. Well that seems not to have worked!
Still getting high peaks.
Still getting high peaks.
Last edit: 3 years 7 months ago by FSM19.
Please Log in or Create an account to join the conversation.
3 years 7 months ago #5693
by FSM19
Replied by FSM19 on topic Gradual count increase at FSM19
Things seem to be better, but for some reason counts are not showing up on Radmon!
Please Log in or Create an account to join the conversation.
3 years 7 months ago #5694
by mw0uzo
Replied by mw0uzo on topic Gradual count increase at FSM19
Is everything working OK now? I can see your graphs starting to update. They redraw at a set interval (except for the one on the map, that is redrawn after every new reading) so give them time to update.
Please Log in or Create an account to join the conversation.
3 years 7 months ago #5695
by FSM19
Replied by FSM19 on topic Gradual count increase at FSM19
Everything is fine now, thanks. I built a new PSU with loads of smoothing, and added ferrites on my probe, ethernet and mains cables. The failure in uploading was due to a faulty network socket.
Please Log in or Create an account to join the conversation.
3 years 7 months ago #5701
by FSM19
Replied by FSM19 on topic Gradual count increase at FSM19
I really don't believe it, everything runs fine for days, and suddenly again more spikes! All I can think of is mains spikes. We are having pretty awful weather at the moment, and in a rural area the power can be inconsistent. Monitoring the other day I found that the nominal 230V varied between 218 and 245!
Please Log in or Create an account to join the conversation.