request for assistance: CPM bounces from actual to 0.
- ConiferCOUSA
- Topic Author
- Offline
- New Member
Less
More
- Posts: 5
- Thank you received: 0
7 years 9 months ago #2964
by ConiferCOUSA
Make/Model: International Medcom RadAlert model 50
Tube: LND712
Hiya. request for assistance. I think it is related to update frequency settings.
I have both set to 60 seconds. (capture and report I think.) My CPM on the live map bounces between what my 1 min average on my geiger counter shows, and 0. The online reporting tool shows the same. Lots of Zeros between my actual raw data. I don't see this behavior on other stations. I must have it misconfigured. Ideas?
Thank you,
Stephen
I have both set to 60 seconds. (capture and report I think.) My CPM on the live map bounces between what my 1 min average on my geiger counter shows, and 0. The online reporting tool shows the same. Lots of Zeros between my actual raw data. I don't see this behavior on other stations. I must have it misconfigured. Ideas?
Thank you,
Stephen
Make/Model: International Medcom RadAlert model 50
Tube: LND712
Please Log in or Create an account to join the conversation.
7 years 9 months ago #2975
by mw0uzo
Replied by mw0uzo on topic request for assistance: CPM bounces from actual to 0.
Ok I've seen this before, I think last time it was because two RadLogs were running at the same time.
Close all RadLogs, check task manager for any hidden RadLogs, close/kill them.
Then start RadLog.
If only one is running, ensure you are using the stable version of RadLog.
Close all RadLogs, check task manager for any hidden RadLogs, close/kill them.
Then start RadLog.
If only one is running, ensure you are using the stable version of RadLog.
The following user(s) said Thank You: nu3e
Please Log in or Create an account to join the conversation.
- ConiferCOUSA
- Topic Author
- Offline
- New Member
Less
More
- Posts: 5
- Thank you received: 0
7 years 9 months ago #2983
by ConiferCOUSA
Make/Model: International Medcom RadAlert model 50
Tube: LND712
Replied by ConiferCOUSA on topic request for assistance: CPM bounces from actual to 0. <<Resolved. Thank you!>>
Wow. You nailed it. There was a second one running. Halted that instance and things look right now. Thank you.
Make/Model: International Medcom RadAlert model 50
Tube: LND712
Please Log in or Create an account to join the conversation.
7 years 7 months ago - 7 years 7 months ago #3197
by Timo70
Replied by Timo70 on topic request for assistance: CPM bounces from actual to 0.
Hi there.
I'v struggled with the same problem for a while.
As you can see in the picture, first i had meny zeros and average was quite low.
I have had changed values of "Serial-audio counting period" to 30s.
Changed it back to 60s, and it seems to be ok, and average peaks up.
What really are the right values, please advise?
I'v struggled with the same problem for a while.
As you can see in the picture, first i had meny zeros and average was quite low.
I have had changed values of "Serial-audio counting period" to 30s.
Changed it back to 60s, and it seems to be ok, and average peaks up.
What really are the right values, please advise?
Last edit: 7 years 7 months ago by Timo70.