Power supplies for modern MCU Geiger counters
1 year 1 month ago #6785
by Simomax
Power supplies for modern MCU Geiger counters was created by Simomax
For quite a while I have been using a good quality 12V 4A switch mode power supply then a 7805 regulator and ancillary components to run my small modern electronic MCU type counters. NetIO/CA JOE/ESPGeiger HW etc. I had to remove that PSU as I need it for something else, but noticed that other SMP's have a hell of a lot of noise or something. I currently have an STS-5 on my breadboard counter and with the original PSU nothing happened if I touched the metal wall of the tube or picked it up. The first PSU I tried as a replacement just made the counter go nuts. Constantly counting, so that went in the bin. The second and third PSUs seem to work OK, but when I touch the tube wall or pick it up, the counter goes nuts, again counting like crazy. Putting the old PSU back on stopped this happening and I could quite happily touch the tube. I have had to use the old PSU for something else, so now when I touch the tube it goes nuts. This is also being regulated by a 7805 so the ripple or interference must be pretty bad.
I highly suspect it is noise from the PSU, and my touching the tube is somehow grounding it or the opposite, but the electrical path is enough to interfere with the counter. I think the way forward is to build an old school linear PSU with some decent capacitors on it. Has anyone else experienced this kind of thing? And if so, what did you do to fix the issue?
I highly suspect it is noise from the PSU, and my touching the tube is somehow grounding it or the opposite, but the electrical path is enough to interfere with the counter. I think the way forward is to build an old school linear PSU with some decent capacitors on it. Has anyone else experienced this kind of thing? And if so, what did you do to fix the issue?
Please Log in or Create an account to join the conversation.
1 year 1 month ago #6786
by Juzzie
Owner and operator of "southofhobart" monitoring stations.
Replied by Juzzie on topic Power supplies for modern MCU Geiger counters
Yes, some power supplies are very noisy. I've had a geiger counter play up by just being near to a bad supply. Where I am there is a "Tip Shop", that is a reuse shop at the waste transfer station/rubbbish dump. There is a fantastic selection of discarded PSUs of all types, and they cost next to nothing, so I can take my pick without financial risk. Not sure if you'd have the same resource...
Maybe you could use your scope and determine the correct capacitor to remedy your problem.
I have tried to add capacitors but to no avail, as it was like a wild stab in the dark without a scope.
Maybe you could use your scope and determine the correct capacitor to remedy your problem.
I have tried to add capacitors but to no avail, as it was like a wild stab in the dark without a scope.
Owner and operator of "southofhobart" monitoring stations.
Please Log in or Create an account to join the conversation.
1 year 1 month ago #6787
by Radslug
Replied by Radslug on topic Power supplies for modern MCU Geiger counters
I've noticed the quality of wall wart PSU's has dropped significantly in the last year or two, if that adds any perspective? Perhaps longer, but it wasn't so severe as now. I run other data collection devices, a few of which are based on the Rock-64 sbc (I grabbed a stash when a 'tech' startup failed). Ordering from Amazon I've learned to order a scattershot of different brands and cases of PSUs that are not obvious rebrands of the same item. Usually 2 or 3 have been unable to hold voltage within the spec'd range of current (or even far below). Right when it seems good quality ones no longer exist outside of Mouser, DigiKey, etc., I get a batch that are rock solid and last for years without issue. I've been too busy to do a proper teardown but I am curious at what has gone so wrong. My obvious guess is that we're going through another time period where entire batches of products from component manufacturers are flawed and either out of spec or can only hold spec until they're under load, and nobody tests anything in the supply chain?
Please Log in or Create an account to join the conversation.
1 year 1 month ago #6788
by Simomax
Replied by Simomax on topic Power supplies for modern MCU Geiger counters
Thanks for your replies. The PSUs I was trying are all used that have come from working environments. Its pretty unknown as to how long they have been in use, but were working when I acquired them as I removed them from equipment I was upgrading. I'm just going to assume they have either had it, or are low quality, high ripple etc. One thing that did come to mind, and I can test this, is one of the polarities at mains, or close to mains potential. I have seen it before where (even on some half decent) either the positive or negative is at mains potential. If that was the case and I am earthed, or even not earthed, and I touch the tube, it could be getting some interference from mains itself. Big Clive has identified this kind of thing more than once.
With regards to the filter capacitor, I wasn't aware that it had to be a certain value, tuned to the PSU. My understanding, (and I doubly checked with chatGPT. It has been known to be wrong, so have I), is that obviously the voltage has to be above that which it is fed, but the value simply has to be big enough to filter the ripple. Increasing the current draw will increase the ripple, so a larger value would be needed in that case. As far as I am aware the value of the capacitor is based on the current that is drawn and the ripple that is created. Some manufacturers may think that 30mv ripple is OK at 2 amps, others may go with 10mv. Depends on what the manufacturer I guess and the intended application for the PSU. I do have (somewhere) a very nice Meanwell, I think, medical grade PSU. The output is completely isolated from mains, completely. It's designed for use with medical devices that may have electrodes or things that attach to the human body like tens machines or something like. It would be interesting to put that on the scope and see what kind of ripple that is giving, and then check some of the cheaper PSUs I tried.
I think the best option for me though is to build a simple linear PSU, just for Geiger counters. I wish you could still buy the old linear PSUs but seems you can't except in certain cases such as high end HiFi, and they are not cheap. I do have a very basic 16v AC 1A transformer wall-wart, so I could easily build a simple rectifier/smoothing circuit and feed my 7805 with that. That would be completely isolated from mains, linear, so no switching ripple.
Right before I post this I just checked the PSU that is running two counters with my scope. It was quick and easy to at least check the 5v rail. This is post 7805. So I'm checking the ripple after the 7805, and it's bad. 30mVPP. It's messy also, as you can see in the picture. I didn't check the 12v side, the actual PSU output, but 30mv ripple after a 7805 is completely unacceptable. I think a linear PSU is really the best way for me.
With regards to the filter capacitor, I wasn't aware that it had to be a certain value, tuned to the PSU. My understanding, (and I doubly checked with chatGPT. It has been known to be wrong, so have I), is that obviously the voltage has to be above that which it is fed, but the value simply has to be big enough to filter the ripple. Increasing the current draw will increase the ripple, so a larger value would be needed in that case. As far as I am aware the value of the capacitor is based on the current that is drawn and the ripple that is created. Some manufacturers may think that 30mv ripple is OK at 2 amps, others may go with 10mv. Depends on what the manufacturer I guess and the intended application for the PSU. I do have (somewhere) a very nice Meanwell, I think, medical grade PSU. The output is completely isolated from mains, completely. It's designed for use with medical devices that may have electrodes or things that attach to the human body like tens machines or something like. It would be interesting to put that on the scope and see what kind of ripple that is giving, and then check some of the cheaper PSUs I tried.
I think the best option for me though is to build a simple linear PSU, just for Geiger counters. I wish you could still buy the old linear PSUs but seems you can't except in certain cases such as high end HiFi, and they are not cheap. I do have a very basic 16v AC 1A transformer wall-wart, so I could easily build a simple rectifier/smoothing circuit and feed my 7805 with that. That would be completely isolated from mains, linear, so no switching ripple.
Right before I post this I just checked the PSU that is running two counters with my scope. It was quick and easy to at least check the 5v rail. This is post 7805. So I'm checking the ripple after the 7805, and it's bad. 30mVPP. It's messy also, as you can see in the picture. I didn't check the 12v side, the actual PSU output, but 30mv ripple after a 7805 is completely unacceptable. I think a linear PSU is really the best way for me.
Attachments:
Please Log in or Create an account to join the conversation.
- DonZalmrol
- Offline
- Senior Member
Less
More
- Posts: 40
- Thank you received: 6
1 year 2 weeks ago #6812
by DonZalmrol
Replied by DonZalmrol on topic Power supplies for modern MCU Geiger counters
Currently I'm using these for my setup(s):
- https://www.rhelectronics.store/high-voltage-400v-geiger-tube-driver-two-channels-with-signal-buffer-for-muon-cosmic-ray-coincidence
- https://www.rhelectronics.store/high-voltage-geiger-probe-driver-module-380v-550v-with-ttl-output
Please Log in or Create an account to join the conversation.
1 year 2 weeks ago #6817
by Simomax
Replied by Simomax on topic Power supplies for modern MCU Geiger counters
I was referring to the 5v PSU that supplies the HV PSU, not the HV PSU itself. The 5v PSU I have is causing some noise in certain circumstances that covers all the electronics on the counter. Two counters in fact as I use that PSU to power more than one Geiger counter.